Power Supply: |
|
|
|
Voltage: |
|
24 VDC (±5%) |
|
|
Current: |
|
1 amp typical; 2 amp peak |
Syringes: |
|
|
Syringe Type: |
|
For 4.5 in module, Hamilton GASTIGHT? Half Height Instrument Syringes |
|
|
Sizes: |
|
12.5 uL, 25 uL, 50 uL, 125 uL, 250 uL, 500 uL, 1.25 mL, 2.5 mL, 5 mL, 12.5 mL |
|
|
Fluid Contact: |
|
Glass, PTFE, CTFE |
Syringe Drive: |
|
|
Drive Type: |
|
Closed loop DC Servo Motor: self lubricating and wear compensating lead nut/screw (no lubrication) |
|
|
Travel/speed: |
|
30mm: 0.5 seconds to 60 minutes for full stroke. Dependent upon system configuration. |
|
|
Resolution: |
|
30,000 increments full stroke on 30 mm stroke length Adjustable to the following counts:
Standard/half |
1,000 counts |
Standard/full |
2,000 counts |
High/half |
15,000 counts |
High/full |
30,000 counts | |
|
|
Accuracy: |
|
±1.0% @ > 30% stroke; ±1.2% @ 5% stroke (syringe dependent) |
|
|
Precision |
|
+0.2% @ > 30% stroke; +0.5% @ 5% stroke (syringe dependent) |
Valves: |
|
|
Valve Type: |
|
Hamilton HV plug valves; snap-in/out |
|
|
Port Size: |
|
.039" (1.0 mm), .059" (1.5 mm), or .070" (1.8mm) |
|
|
Positions: |
|
2 position @ 90° typical; multiple position possible (15° increments) |
|
|
Fluid Contact: |
|
PTFE & CTFE |
|
|
Fittings: |
|
Hamilton 1/4" - 28 male tube fittings with flared FEP or PTFE tubing |
Valve Drive: |
|
|
Drive Type: |
|
Closed loop DC motor with encoder |
|
|
Speed: |
|
390 ms @ 90° |
Environment |
|
|
Operating Temperature: |
|
60°F to 95°F (15°C to 35°C) |
|
|
Operating Humidity: |
|
20% to 90%, non-condensing |
|
|
Storage Temperature: |
|
-5°F to 158°F (-20°C to 70°C) |
|
|
Storage Humidity: |
|
20% to 95%, non-condensing |
Communication |
|
|
Protocol: |
|
Hamilton Reno Protocol 1 or Hamilton Bonaduz DIN Protocol |
|
|
Interface: |
|
RS-232C |
|
|
Baud Rate: |
|
9600 |
|
|
Daisy Chain: |
|
16 modules |
Software Features: |
|
|
Memory Storage: |
|
FLASH |
|
|
Stored Methods: |
|
1 |
|
|
TTL I/O: |
|
1 |
|
|
Auxiliary Input: |
|
1 (e.g. hand probe) |
|
|
Step/Prime Modes: |
|
Yes |
|
|
Firmware Revision: |
|
Programmable FLASH (In-System) |
|
|
Syringe Program Variables: |
|
Position, Speed, Prime Speed (system default), Backlash Compensation (return steps), & On The Fly Speed Change |
|
|
Syringe Program Variables: |
|
Delays Overload/Zero Sensitivity TBD through testing |
|
|
Valve Program Variables: |
|