STANDARDS & REGULATORY WATCH
Biomedical Instrumentation &Technology 483
Don’t: Substitute for BIs | Do: Document Everything | Don’t: Flash Sterilize Implants
Steam Sterilization Do’s and Don’ts
Saturated steam under pressure is one of the oldest
methods used in healthcare facilities to sterilize
medical devices. A new guidance document from
AAMI—Comprehensive guide for steam sterilization and
sterility assurance in health care facilities (ANSI/AAMI
ST79:2006)—provides hospitals with guidance on every
aspect of steam sterilization, from facility design to personnel
considerations, through cleaning, decontamination,
and sterilization, and concluding with quality
assurance. This month, BI&T will examine the guidance
ST79 provides in some key areas concerning steam sterilization
safety and the sterilization of implantables:
1. Class 5 integrating indicators or enzyme-only indicators
are not a replacement for biological indicators
for loads containing implants.
2. Proper monitoring and documentation are essential
to any sterilization and infection control program, and
critical in cases where implants are released before the
biological indicator test results are available.
3. Flash sterilization should not be used for implantable
medical devices.
Biological Indicators
The use of biological indicators (BIs) provides evidence
of efficacy of a sterilization process by challenging the
sterilizer with a large number of highly resistant bacterial
spores. Biological monitoring provides the only direct
measure of the lethality of a sterilizer cycle. Sterilizer
makers validate their cycles using BIs.
While the performance of Class 5 integrating chemical
indicators (CIs) and enzyme-only indicators has been
correlated to the performance of BIs, these sterilization
monitoring devices do not contain spores and thus do
not directly measure the lethality of a sterilization cycle.
They can, however, provide additional information
about the attainment of the critical parameters of the
sterilization process.
BIs should be used with a user-assembled or commercially
available disposable or preassembled process challenge
device (PCD):
to routinely monitor sterilizers at least weekly, but
preferably every day that the sterilizer is in use;
for sterilizer qualification testing after installation,
relocation, malfunctions, major repairs, and sterilization
process failures; and
to monitor every load containing implants.
Every sterilization load containing implants should
be monitored with a PCD containing a BI (a BI challenge
test pack). A Class 5 integrating CI or an enzymeonly
indicator should be included in this PCD.
ST79 bottom line: With respect to implantable
devices, biological monitoring is necessary to provide
optimal sterility assurance. A Class 5 CI or an enzymeonly
indicator should be included with the BI in the
PCD so that if an implant must be released on an emergency
basis, additional information about the critical
parameters of the sterilization process will be available
and documented.
Kurt Larrick
Kurt Larrick is AAMI’s director of technical publishing and
marketing.
ANSI/AAMI ST79:2006
Implantables should not be flash sterilized.
Every sterilization load containing implants
should be monitored with a PCD containing
a BI.
Written documentation defining emergency
situations in which an implant may be released
before the biological indicator is available
should be developed.
Implant/implantable device: According to FDA, a
“device that is placed into a surgically or naturally
formed cavity of the human body if it is intended to
remain there for a period of 30 days or more. FDA
may, in order to protect public health, determine
that devices placed in subjects for shorter periods are
also ‘implants.’” [21 CFR 812.3(d)]. (From
ANSI/AAMI ST79:2006)
Load Monitoring, Release,
and Documentation
Physical monitoring provides real-time assessment of
the sterilization cycle conditions and provides permanent
records by means of chart recordings or digital
printouts. Physical monitoring is needed to detect malfunctions
as soon as possible, so that appropriate corrective
actions can be taken.
Every sterilization load should be physically monitored.
Every packaged item should be labeled externally
with a process indicator (Class 1 CI) and should contain
an internal CI (Class 3 CI, Class 4 CI, Class 5 CI, or
enzyme-only indicator). If desired, a PCD containing a
BI (a BI challenge test pack) or a PCD containing either
a Class 5 integrating CI or an enzyme-only indicator (a
CI challenge test pack) may be placed in the area of the
chamber and load considered least favorable to sterilization.
The PCD should be equivalent to the BI challenge
test pack described in 10.7.2.1 of ST79. As recommended
above, every sterilization load containing implants
should be monitored with a PCD containing a BI.
Ideally, every reprocessed medical device, especially
an implant, should be traceable to the patient on whom
it is used or in whom it is implanted, and any load that
includes implants should be quarantined until the results
of the BI testing are available. Written documentation
defining emergency situations in which an implant may
be released before the biological indicator is available
should be developed in consultation with infection prevention
and control, the surgeon, and risk management.
When documented medical exceptions dictate (e.g.,
the need for trauma-related orthopedic screw-plate
sets), it could be necessary to release an implantable
device before the BI results are known. In this case,
according to ANSI/AAMI ST79:
the release of the device before the BI results are
known should be documented; and
the BI result obtained later should also be documented.
Releasing implants before the BI results are known is
unacceptable except in limited, emergency situations.
Emergency situations should be defined in written guidance
developed in consultation with infection prevention
and control, the surgeon, and risk management.
Steps should be taken to reduce the frequency of emergency
release of implantable items. For example, ongoing
periodic reviews of the exception forms and implant logs
could reveal consistent patterns of events that are causing
emergency release and that could be corrected. (See
ST79 for examples of an implant log and exception
form.)
Flash Sterilization of Implants
Sterilization technology, even with its steady flow of
advances and new equipment, does not solve the problems
of using flash sterilization for implants. Concerns
about aseptic transfer remain, especially if the sterilizer
does not open directly into the room containing the sterile
field where the device will be used or into an area
either within or directly adjacent to the procedure room.
Careful planning, appropriate packaging (e.g., packaging
that allows the user to see the device for sizing and
verification of features), and inventory management in
cooperation with suppliers can eliminate the need to
flash sterilize implantable items. This is a goal that all
institutions should strive to achieve. For more information,
refer to ANSI/AAMI ST79:2006.
Steam Sterilization Do’s and Don’ts
STANDARDS & REGULATORY WATCH
484 November/December 2006
ANSI/AAMI ST79:2006, Comprehensive guide
to steam sterilization and sterility assurance in health
care facilities. Order code: ST79 or ST79-PDF.
List/Member price: $200/$100.
Sterilization in Health Care Facilities,
2006-2007 edition. The new edition features several
important additions to this essential hospital
resource. Includes 15 AAMI documents; 1,046
pages. Order code: STBK06-1. List/Member price:
$350/$250.
AAMI Standards on CD-Sterilization Edition.
The ultimate resource for sterilization professionals
has been updated to include new material. Now
features more than 50 current AAMI standards and
guidance documents for sterilization on one CD.
Order code: STBK-CD. List/Member price:
$655/$475.